1. RE: BAMs
Feb 2 2011, 1:41 PM EST
We just recieved our BAM PM10-25 or PMc system. (tow units hooked together gives out PM10, PM2.5 and PMcoarse.I am having issue with the data import templete as Bar. Pressure is not coming back. Do you find this valuable? |
|
NSCAPCD |
2. RE: BAMs
Mar 2 2011, 2:09 PM EST
Bob, Sorry to take so long to answer. Although I'm not intimately familiar with the combo BAMS, the feed for BP comes from the MetOne BP/AT BX-596 Ambient AT/BP Combo Sensor that is typically wired to Channles 6 and 7 on the back of the BAMs. This signal is typically not stored or logged by the BAM its simply used internally to adjust to the various sample condtions. To get external BP logged by the BAMs for collection I jumpered a wire from the sensor BP + input channel to the + side of one of the 6 programmable BAM logging channels on that BAM. Once you know the channel has the BP jumpered to it you simply have to indentify(count) that data point position in the data string output by the BAM and that is your Column # for your file import column mapping. Hope this helps I sent you a PM also.